(Topic ID: 288331)
Gottlieb Mars - Correcting a Mistake, and TP Voltages
By Knxwledge
  • 35 posts
  • 4 Pinsiders participating
  • Latest reply 3 years ago by Knxwledge
#1 3 years ago
Before moving on and plugging in the rest of the boards to my Mars, I want to make sure on a couple of things. I've rebuilt the mandatory parts on the power supply board according to Pinwiki, repinned 3 molex plugs that plug in to power supply, done the backbox ground mods (but not ran anything from the head to the bottom of the cabinet or messed with the interconnect harness), replaced the BR and caps on the sound/speech power supply, changed out the 3 BRs and the original orange Sprague cap in the bottom of the cabinet, and made sure all fuses are rated appropriately. I changed the orange Sprague with a 12000uf 25v Ritchey, the sound/speech cap with a 4700uf 63v Kemet, and all BRs with 400v 35a. Parts were bought from Digikey and GPE.
The mistake I made was wiring the BR on the sound/speech board wrong initially. I goofed, and since there was no clear indicator for pos and neg on the old BR, I decided not to look in the manual and assume I knew. Bad idea, but not catastrophic. When I rebuilt everything and powered the game on for the first time, after about 30 seconds the small 47uf cap blew. Pulled the board out, checked the diode and resistor in that area which were fine, then realized I had mounted the BR with negative where positive was supposed to be, and therefore soldered all lugs wrong. I've now soldered the lugs correctly, but is there anything else I could have damaged on that board? I would like to check before I plug it in again.
Also, I'm getting these voltages from the power supply board (with nothing else plugged in):
TP1: 68v
TP2: 50.6v
TP4: 5.1v adjusted with the pot (should I turn it down closer to 5.0v?)
TP5: 6.95v
Now I notice that TP1, 2 and 5 close, but not exact. Is this fine? I'm assuming yes because they're not under load/potentially could be unregulated but I want to confirm first. I don't have Fuse F3 (8a slow for controlled lamps) because I don't have any on hand, or any close to it like a 7a or 7.5a slow that I could just stick in for testing purposes. Doubt that makes a difference.
#2 3 years ago
One more question... is the wire that goes to the chaser light board from the A2J3 housing I repinned in the correct spot, or should it be in the pin above it? I cut it and wasnt paying attention, and I couldnt figure it out from the schematics
#3 3 years ago
I found this video of a guy testing a System 80 under no load and his voltages were closer to what they were written as on the board than mine.
Guess I should start troubleshooting?
#4 3 years ago
All measurements are in DC
J1 pin 1: 9.5v (labeled as 12vdc source)
Banded side J2: 83.6v (labeled as 76.4v)
Ic1 pin 10: 5.7v (labeled as 7.0v)
J1 pin 7: 41.3v (labeled as 60v AC input)
So Im thinking I either have a problem with the way I wired the BRs in the cabinet or a problem with the transistors on the power supply board (Q1, Q2 and Q3)? Likely not Q3 because my 5v is good and my 12v appears to be low directly where it enters the board. Any help appreciated
Edit: I was reading this thread https://pinside.com/pinball/forum/topic/gottlieb-system-80-power-supply-advice#post-6013748 and I see talks of CR7 causing issues on the 8v. I will check this, because I've already replaced the resistor R10 with the suggested upgrade.
#5 3 years ago
Had some time tonight, let me report my findings. Very strange.........
So I checked Q1 Q2, R1-R5, CR5 and CR6. Everything was good, but I replaced Q1 Q2 anyway. A couple of the BR wires in the bottom of the cabinet were crimped, not soldered, so I soldered them to ensure a good connection. I then checked every output on the transformer because that's the last thing it could be, really... Attached is a picture of my findings, and I'm sure my handwriting is chicken scratch to most, so I will write down the strangest findings. All are in AC:
8v (blk/orn) = 16.2v
25v (red/slate) = 44v
60v (lug 9) = 88v
10v (lug 16) = 35v
10v (lug 17) = 45v
12v (lug 7) = 23v
It's looking like bad transformers! Unless someone has a way to test them differently to make 100% sure, it's looking like I'm gonna need to find some kind of replacement or buy good used ones. The fuses in the bottom of the cabinet aren't blowing.
#6 3 years ago
Quoted from Knxwledge:
Just passing through.
Looks like you measured the transformer voltages with your black meter lead on ground?
AC voltages out of the transformer need to be measured across the transformer windings, i.e one multi-meter lead on one side of the winding, the other multi-meter lead on the other side of the winding.
So, to measure the 10VAC output, one meter lead goes on transformer lug 3, the other meter lead on lug 7.
60VAC across lugs 8 and 9
12VAC across lugs 16 and 17
The centre tap windings at lugs 10, 11, 12 actually has DC applied to the centre lug 11 downstream at one of the circuit boards. Measure the DC voltage at lug 11 with the red meter lead, black meter lead on ground.
Then measure the AC voltages between lugs 10 and 11, then 11 and 12.
Ditto for the centre tap windings at lugs 13, 14, 15 respectively. Lug 14 also has DC applied to it downstream.
#7 3 years ago
Quoted from Quench:
Thanks, I will retest correctly and report back
#8 3 years ago
Quoted from Quench:
Transformer voltages:
10vac (lug 17 and 16) is 13.18vac
60vac (lug 8 and 9) is 63.97vac
Lug 11 dc voltage is 1.0 but inconsistent
Ac from 10 to 11 is 2.8vac
Ac from 11 to 12 is 2.8vac
Lug 14 dc voltage is 2v but inconsistent
Lug 13 to 14 ac is 1.6vac
Lug 14 to 15 ac is 1.6vac
Larger transformer:
Ac at 8v lugs is 9.09vac
Ac at 6v lugs is 6.9vac
The inconsistent DC voltages started higher and as the cabinet was turned on, it went down. Seemed to stabilize at the voltages I posted
Dc voltages at the fuses with black lead on cabinet ground:
F1: readings start at 310v and go down to 275v
F2: 279v
F3: 7.7v
F4: starts at 20v and goes down to 0-0.2v (with sound speech PS unplugged)
F5: 24v
F6: 48v
F7: starts at 7v and goes down to 0-0.2v
Just cleaned the fuse holders, no change
Dc voltages on the BRs (with black lead on negative, red leads going clockwise):
Power supply/sound board lug: 23.5v/28.3v/24v
Lamp: 1.4v/20v/8.5v
Solenoid: 48v/80v/47v
I checked the rest of the resistors on the power supply board and the only thing I found bad was R12. It's a 0.33ohm 5w wirewound and it is reading 0.7ohm. Do the 1000pf and 2uf caps at C3 and C5 ever go bad? Can I replace a wirewound resistor with a metal film one?
Checking Q3:
One leg to another 0.6
Swapped leads 2.0
Black lead on legs, red on base tested open both times
Red leads on legs, black on base tested 0.6 both times
#9 3 years ago
To answer my own question in regard to the miswiring of the sound/speech BR... the other 2 lugs that arent positive and negative dont lead anywhere except for the molex plug. So no, I didnt damage anything else besides that one cap which blew.
Anyone have any advice on the power issue? Leaning toward a bad IC1, but I dont think that would explain the low 12v
#10 3 years ago
The voltages at the fuses are on the AC side before the bridge rectifiers, measuring them with respect to ground on DC isn't how to test them.
You can check them with the power on by setting your multi-meter to AC voltage and put a meter lead on each side of the fuse. You should read zero volts.
Better yet, measure AC voltage at the fuse with respect to the opposite winding lug at the transformer.
When you put your meter on resistance mode and put both probes together, do you read zero ohms? probably not. The discrepancy on that R12 resistor is likely just your meter not being calibrated. When wirewound resistors fail, they usually go open circuit.
Which 47uF capacitor blew?
To measure the DC voltages at the bridges, set the multi-meter to DC voltage, black meter lead goes on the "-" bridge leg, red meter lead goes on the "+" bridge leg.
#11 3 years ago
Quoted from Quench:
The cap that blew was on the sound/speech PS board because I installed the BR backwards. It's fine now.
The resistance between my 2 meters is about 1.3ohm. So I guess the one on the PS is fine
I have another wirewound resistor of the same rating under the PF that needs to be replaced. It's testing 2.4ohm. Is the metal film one fine to use? Is the second reading I got when testing Q3 suspect, or is it okay?
I will retest the fuse AC voltages (with the exception of F3 because I don't have the 8 amp SB fuse at the moment)
Fuses tested on AC:
F1: 0
F2: ~55
F3: cant test
F4: 6.3
F5: 0
F6: 0
F7: 0
I checked the fuses, theyre fine, except the 7.5a one at F2 is an 8A one. I checked the 115v on both transformers, and with 1 lead on 115, 1 lead on common I got 117vac for both.
I rechecked the BR DC measurements and Im getting:
Power supply/sound board: 16.7v
Lamps; 4.0
Solenoids: 35.5v
#12 3 years ago
Some pictures (I know R12 isnt soldered in/the metal film one instead of the wirewound)
20210224_183355 (resized).jpg20210224_183355 (resized).jpg20210224_183358 (resized).jpg20210224_183358 (resized).jpg20210224_183408 (resized).jpg20210224_183408 (resized).jpg20210224_183413 (resized).jpg20210224_183413 (resized).jpg
20210224_183736 (resized).jpg20210224_183736 (resized).jpg20210224_183733 (resized).jpg20210224_183733 (resized).jpg20210224_183739 (resized).jpg20210224_183739 (resized).jpg
#13 3 years ago
I don't see any wirewound resistors on the playfield schematics (maybe I just missed it). Where is this resistor? Is it also supposed to be a 0.33 ohms? Unsolder one leg and remeasure it to get its real resistance value so the reading isn't affected by other circuitry.
If you're going to replace wirewound resistors with metal film, be very careful to get the wattage rating same or higher on the replacement. Note ceramics are generally larger in size so are better capable of dissipating the heat.
Fuses tested on AC:
F2: ~55
F4: 6.3
Well those readings indicate fuse F2 is open circuit and F4 is suspect - or there's bad connections on the fuse holders.
Your opening post says you've got 5.1V at TP4 on the power-supply board which is the output of Q3 and is about correct. Is this not the case anymore?
9.5V sounds low to me which indicates the big capacitor across the 12V bridge isn't doing its job.
BTW, what exactly is the issue you're trying to sort out right now?
#14 3 years ago
Quoted from Quench:
I got the 2.4ohm reading with one leg unsoldered. I will take a pic of where it's located under the PF
Quoted from Quench:
I will try swapping the fuses, and if that doesn't change anything I could try changing the fuse holder block.
Quoted from Quench:
No, my 5v is fine. Is Q3 only for the 5v?
Quoted from Quench:
The high 60v, 42v and low 8v indicated in post 1
TP1: 68v
TP2: 50.6v
TP5: 6.95v
This big capacitor, are you referring to the 470uf cap on the PS board? I've changed that
#15 3 years ago
There are several readings from earlier that are confusing. I think many are due to how they are being measured.
But based on latest info:
If TP4 is correct then don't mess with the 5V regulator circuit including Q3.
Note that he is not a standard transistor, he is a "Darlington Pair" transistor which has two transistors in the package. These do not test the same as normal transistors do.
TP1 -- you need to measure the 'setting voltage' which can be found at location "E2".
Meter grounded to TP3 - what voltage do you get at "E2".
It is a little high but not horrible. About 10% out of whack when it should be closer to the 5% range.
TP2 is fine. When you fix TP1 then TP2 will fall in line.
TP2 is calculated by voltage at TP1 minus 18V. You normally get about 60 - 18 = 42V. You get 68 - 18 = 50V. This says CR6 and R5 are good.
TP5 is bad. Replace CR7 and R10.
Out of curiosity: meter grounded to TP3 - what voltage do you get at "E4"?
#16 3 years ago
this may not help, but when getting my Solar Ride system 1 up and running i noticed that when testing voltages in the cabinet, on the base board, that the voltages were incorrect until it was all connected and there was load for the circuit in question.
#17 3 years ago
Quoted from G-P-E:
Before I go through the trouble of putting the PS board back to test E4 and E2, I have already replaced CR7 and R10 with the suggested "upgrades" (I will double check I used the right resistor). I replaced the electrolytic caps on it, replaced the 500ohm pot, added a ground wire, replaced Q1 and Q2 despite them testing fine, changed out the 3 male molex pins, repinned the female pins and checked every single resistor and diode on the board. The only things I haven't really checked are SRC1 and IC1
#18 3 years ago
Quoted from G-P-E:
Quoted from Quench:
Picture of the location of the 0.33ohm 5w resistor on the pf. The other leg is connected to the cap and the unsoldered wire.
Changed fuses with known good ones, nothing changed.
I actually found a 7a 250v SB fuse to test F3 and it tested good
For R10 I used a 680ohm 1/2w resistor, and a 8.2v 1w zener for Cr7
#19 3 years ago
SCR1 and IC1 have nothing to do with the 8V. These are part of your 5V supply which is working.
ONLY CR7 and R10 affect the 8V offset back to the transformer.
If you have a bad solder joint or cracked trace - that can be an issue. So can a mis-pinned plug on J1.
With meter grounded to TP3. Measure your voltage at E2 (for 62V supply) and E3 (for 8V supply).
#20 3 years ago
Quoted from G-P-E:
No cracked joints. I verified I repinned J1 correctly
All measurements in Dc:
E2: 69.2v
E3: 6.9v
E4: 9.5v
Are these BR measurements a problem(the ones in the cabinet next to the filter cap)?
Lamps; 4.0
Solenoids: 35.5v
#21 3 years ago
E2 is high, E3 is low.
E2 is established by CR5 and R1.
If resistance in R1 goes too high, there will be insufficient current and CR5 won't conduct.
Check solder job on R1 and CR5 again. Check R1. Also possible CR5 can be damaged during installation.
E3 is low and may be due to low incoming 12V. With a 680 ohm resistor on input, the voltage drop across him my be too great and not allow the zener to hit knee voltage. We need to get voltage at E4 up before worrying about E3.
First step is to follow the voltage back to the source.
Going by your "Power supply/sound board: 16.7v" mentioned in last post. Assuming they share the same unregulated power - this says there is 16.7V in base but 9.5 at power supply. Screams bad connection if true. Voltage at E4 is labeled 12V DC but can often go a few volts higher (e.g. 16.7) with no problems.
At the black cap with screw terminals as shown above -- measure the voltage directly across the screw terminals.
One thing at a time - worry about lamp and solenoid voltages after you get the main 5V and display power up.
#22 3 years ago
Quoted from G-P-E:
When removing Q3 again I think I may have broke it. If the leg moves up and down in the case, Im gonna guess thats not okay
R1 gives me 1.3k, Cr5 gives me 0.699 and open using diode test, swapping leads. Both were done with 1 leg out of circuit. Solder joints on Cr5 and R1 look fine
Should I change R10 to something else?
Measurements at filter cap:
Dc is 571v
Ac is 0v
#23 3 years ago
that's the biggest downfall of the gottlieb power - having to remove the transistor to do any other work.
571V? Huh?
What are you using for a meter?
Don't worry about R10 right now.
#24 3 years ago
Quoted from G-P-E:
571V? Huh?
Im using an Aneng AN8009, I grabbed another meter, a Sperry Dm-6400, and Im getting 0.57v DC, 0v AC with it. Just tested it with the Aneng again, and now Im getting 24v DC and the Sperry says 0.024v DC.... I dunno what the deal is with these meters
#25 3 years ago
Ignore everything in post #24, I figured it out and I'm not happy with myself. It turns out BOTH crimps on the filter cap were bad. I yanked the wire right out of both terminals with little effort. Recrimped them properly, and I'm getting 14.7v at the filter cap, the 12v BR and J1 pin 1. I'm relieved, but also very annoyed with myself
Im still having the same issues with my F2 and F4 fuses, and the 6v Lamp bridge tests 12v, and the 24v Solenoid bridge tests 36v. My F8 primary fuse also gives me ~70v AC
#26 3 years ago
https://homepinballrepair.com/finding-and-testing-fuses/
The only thing I found on what to do if you have AC current on the fuse block is to replace the fuse... but Ive already tried that and it didn't change anything. I will try changing the primary fuse and see if it affects the AC current on the other two
Edit: Changing the fuse didn't work. The primary fuse F8 now tests ~26v AC tested from one edge of the holder to the other.
#27 3 years ago
A fuse is a dead short and you should never see a significant voltage drop across it.
If you have a voltage drop across a fuse holder then the fuse is bad or the fuse's connection to the fuse holder is bad.
What brand of fuse did you put in it? Is the fuse holder tight? Or is it too easy to pull/reinsert the fuse?
And are you sure the meter is saying 26VAC? Or maybe 26mVAC (0.026VAC)?
#28 3 years ago
Quoted from G-P-E:
This ended up being it. I didn't realize in the bottom right of my auto-ranging meter it said mV. All readings I got from the fuses were in mV not V. Thank you
I think the last question I have for the moment is, am I good to replace the 0.33ohm 5w wirewound resistor under the PF with a metal film one of the exact same rating? Picture in post 18. Other than that, I've gotta order another Q3 (which are $11 at Marcos!!!!!), reinstall the PS board and make sure my voltages are fine once more, then power it on for real
#29 3 years ago
mV makes sense -- you're ok with low readings in the mV range.
That resistor I don't know.
For where he is located - I would say as long as you match the wattage and resistance then you should be fine.
For the transistor - you can do better.
2N6059 and 2N6284 are better than original replacements for the 2N6057.
Make sure to reapply new heat sink compound!
#30 3 years ago
Quoted from G-P-E:
Ooooh okay I should have those, I was confused about that. What can I replace these with? Theyre under the PF as well, should I just swap them even if theyre not bad?
20210227_112533 (resized).jpg20210227_112533 (resized).jpg20210227_112544 (resized).jpg20210227_112544 (resized).jpg
#31 3 years ago
If they work - leave them alone. They can last decades longer.
#32 3 years ago
Quoted from G-P-E:
Great, will do. Thanks for your help so far, Im sure I will need more when I get to turning the game on fully, but this should get me going for quite a while.
#33 3 years ago
Quoted from G-P-E:
Scratch my last message, something weird is going on. The filter cap is holding 10.95v DC with the game unplugged, and when I plugged in J1 on the power supply the LEDs lit up right away. I instantly unplugged it. This is all with the game completely powered off. Is this normal? Maybe when I unplugged the terminals from their poorly crimped housing I shorted the cap or something. The 10.95v reading went up to ~14v once I powered the game on, then turned it off again
With black lead on pos lug and red lead on neg lug Im getting a short on continuity test on the filter capacitor. Wires still connected
My 24v solenoid bridge is testing 77v now, and my 6v lamp bridge is testing 11 now. I get the feeling these readings arent accurate, but Ive checked it with 2 meters and I get the same thing
#34 3 years ago
Think of a capacitor as a short term battery. In an overly simplified explanation, it charges up to a voltage while machine is turned on and drains down when machine is turned off. While turned off, the new cap is holding the charge long enough to light the LED but will discharge quicker as you attach other loads to the power supply. This shows the cap is working properly. Don't worry about this - it's normal.
For the solenoid and lamp voltages - DMMs often give phantom readings on low load sources. Only way to get a sure measurement is using an OLD needle type meter.
#35 3 years ago
Quoted from G-P-E:
Got it, I didnt notice this same behavior when I was working on my Bally Lost World, so I wasnt sure if it was normal. Thank you!
$ 12.00
Tools
Nezzy's Pinball Prints
$ 83.00
PinballReplacementParts
This page was printed from https://pinside.com/pinball/forum/topic/gottlieb-mars-correcting-a-mistake-and-tp-voltages and we tried optimising it for printing. Some page elements may have been deliberately hidden.
